Conductivity Calibration

Question on the conductivity calibration procedure for the Aquacontroller Pro.

Page 31 of my manual, Step 2 states:

Select the appropriate temperature compensation value. Most solutions have a temperature compensation correction factor of 2.0 to 2.5 %/C.

I'm not sure what that means. I have this calibration fluid:

Calitech Conductivity Standard.
447 uS/cm +/- 1.0% @ 25*C


So what value should I enter to the Aquacontroller?
 
It does. That is what that temperature compensation factor is for. Salt water has a temperature correction factor of 2.1 to 2.2 %/C.

Curt
 
OK, so the value to enter does not depend on the testing solution? The wording on the instructions made it sound like it might be some number on the packet (which I couldn't find).
 