I've just performed the tests on some ORP calibration solutions. Here's the story:
Over night I cleaned my ORP probe using dilute bleach to remove organic deposits followed by dilute Hcl to remove calcium deposits. I rinsed the probe well in DI water.
Today, I carried out some tests of my ORP calibration solutions. First, I mixed the YSI Zobell Solution (3682) using 125ml of de-ionized water (Hanna) as instructed. At 25C (the weather has been nice in England yesterday and today
) the Zobell reference solution claims 231mV +/- 10mV.
For each solution I rinsed the ORP probe in de-ionized water and then in the calibration solution, before placing the probe in a clean 10ml vial filled with the 230mV calibration solution.
I calibrated the ORP probe of my IKS aqua-computer using the fresh Zobell solution as a reference standard. After calibration I took a reading and the meter indicated 231mV.
I then tested the following samples of IKS 230mV redox calibration solution:
Sample 1
Date of purchase: 4/8/03
Opened: Never
Valid to: 08/03
Reading: 215mV
Sample 2
Date of purchase: 10/10/02
Opened: Once
Valid to: Not given on bottle
Reading: 236mV
Sample 3
Date of purchase: 10/10/02
Opened: Twice
Valid to: Not given on bottle
Reading: 212mV
Given the precision of +/- 10mV on the Zobells I freshly mixed and also the high temperature dependence of the Zobell I think the measurements were all within the margins of error.
I feel much more confident of these solutions now!
I've just performed the tests on some ORP calibration solutions. Here's the story:
Over night I cleaned my ORP probe using dilute bleach to remove organic deposits followed by dilute Hcl to remove calcium deposits. I rinsed the probe well in DI water.
Today, I carried out some tests of my ORP calibration solutions. First, I mixed the YSI Zobell Solution (3682) using 125ml of de-ionized water (Hanna) as instructed. At 25C (the weather has been nice in England yesterday and today

For each solution I rinsed the ORP probe in de-ionized water and then in the calibration solution, before placing the probe in a clean 10ml vial filled with the 230mV calibration solution.
I calibrated the ORP probe of my IKS aqua-computer using the fresh Zobell solution as a reference standard. After calibration I took a reading and the meter indicated 231mV.
I then tested the following samples of IKS 230mV redox calibration solution:
Sample 1
Date of purchase: 4/8/03
Opened: Never
Valid to: 08/03
Reading: 215mV
Sample 2
Date of purchase: 10/10/02
Opened: Once
Valid to: Not given on bottle
Reading: 236mV
Sample 3
Date of purchase: 10/10/02
Opened: Twice
Valid to: Not given on bottle
Reading: 212mV
Given the precision of +/- 10mV on the Zobells I freshly mixed and also the high temperature dependence of the Zobell I think the measurements were all within the margins of error.
I feel much more confident of these solutions now!