Want more accuracy from your test kits???? Read this to find out how.

disc1

-RT * ln(k)
There are a ton of complaints about the accuracy of the kits we use to test our tank water. Many of our kits are highly precise though. What I mean by that is you get the same result from the same kit, even though two different kits may not agree. Even between two kits from the same brand, the results often vary by a good percentage.

So what can you do if you want a more accurate result? If you are using one of the titration based test kits, you can standardize the kit yourself. I'll use for my example here a Salifert kit since that is what I have on hand, but the same procedure will work for any of the titration tests, those where you add a reagent from a syringe and measure the amount delivered to get the result.

First let's talk about how these tests work and then we'll see how to standardize a kit. In a titration experiment, you are adding a reagent with a known concentration to one that is unknown. Let's take the example of a simple magnesium titration. The titrant will be a chemical called EDTA. Every molecule of EDTA will react with one magnesium ion in a reaction called chelation that binds the two together. The details aren't important, just that ratio, 1 EDTA reacts with 1 magnesium ion. We also have an indicator dye. It's called Eriochrome Black T. When it is in solution with magnesium, it binds to the magnesium and that gives it a pink color. When there is no magnesium, it has a bluish color.

So when we start, we add some of the indicator to the sample. We also buffer the sample to a pH of around 10 but that's not terribly important here. The solution turns pink because there is enough magnesium around to bind with the indicator.

Now we start adding the EDTA solution. We know exactly the concentration of this solution. We add the EDTA solution slowly. Now EDTA binds to magnesium better than the indicator does. So EDTA can remove the magnesium from the indicator, turning it blue. At first, this doesn't happen because there is more magnesium floating around in solution. When we have added exactly the same number of EDTA molecules as there are magnesium ions in the solution, then even the magnesium that's bound to the indicator gets chelated away by the EDTA. This turns the indicator blue.

Now since we know exactly the concentration of our EDTA solution, and we know the volume that we used, we know exactly how many molecules of EDTA we added. Since each molecule of EDTA react with exactly one magnesium ion, that tells us how many magnesium ions are in the sample.



But what if we don't know exactly the concentration of our EDTA solution? That's the predicament we are in with a test kit. It's been in that bottle for a while, and who knows what was going on at the factory that day. Things are probably not going to be exact.

So what we need to do is standardize it. We are going to use a sample with a known concentration and do that titration math backwards to calculate the exact concentration of our titrant.






So let's see how to do this with a Salifert kit. The same procedure should work with any titration type kit.

The IAPSO standard has a magnesium concentration of 54.0mM. That equates to ( 54.0mM * 24.305 mg/mmol = ) 1312.47 ppm. That's the number I'll use here, but you can use whatever standard you have.

Now let's run the normal test with the salifert magnesium kit on the standard solution. The first two reagents are the pH buffer and the indicator. They're going to be added in excess, and their measurement does not affect the result we are going to get. Now we do the titration. We fill the syringe and make note of the position, either with the tip of the black piece or the top of the liquid or whatever point is easiest for you to read. It doesn't matter because we are only considering a difference. It's easiest to do the math if you make this the 1.00mL mark, but you don't have to. Just make sure there's no little bubble in the tip. Now add drop-wise with stirring until we get the color change. At that point the titration is over. Read the syringe using the same point of reference you used before (tip, top of liquid, or whatever). Subtract the two numbers and you now know how much titrant you used.

Let's say we marked at the 1.00ml mark before and at 0.15 ml mark at the end. So we used ( 1.00 - 0.15 = ) 0.85mL of titrant. (Notice if we started at 0.90 we would have ended at 0.05 and ended up with the same 0.85ml.)

Now here's the trick of the math. We are doing the same math we are going to have to do later when we test our tank water, only backwards. So we don't need to go all the way to figuring out the actual number of molecules. Trust me on this, or don't. I can write you a proof on it if you need me to.

Now we know that the number of magnesium ions in our sample is equal to the number of molecules of the titrant. So ( 1312.47 ppm Mg / 0.85ml titrant ) = 1544.08 ppm Mg / ml titrant. I know those units look weird, but if you watch the math later, their going to work out in our favor.

Now look at your Salifert instructions. Running 1ml of titrant would be a result of 1500 ppm, where in our case it would be 1544 ppm. We have discovered the size of the error. We can also see that we measured a 1312 ppm standard and got a result of 1275 ppm. Notice that the size of the error isn't constant. It gets larger the larger your measured value gets in a linear fashion.

Let's save this number 1544. Write it down on the top of the box so you have it in the future when you are running your tests. And let's test our tank water. We run the test just like we did on the standard and let's say we find that we used 0.78 ml of titrant.

We multiply that by the number we got in our standardization to get ( 0.78 ml titrant * 1544.08 ppm Mg / ml titrant = ) 1204.3824 ppm. OK, no teacher is going to let you keep all those significant figures. I'll let you keep out to the tens digit. I think that is safe given the way we did our math. That means we got 1200, but we know for certain now that the real value of our magnesium concentration in our tank is between 1195 and 1205.

The shortcuts we took on the math are actually working for us. By excluding sample volume, it doesn't matter what volume you use so long as you use the same volume for standardization as you do for analysis. We would never do this in the lab, we would get the concentration, but here it's going to protect us from the error in measuring our sample with that syringe. Those things are horribly inaccurate and I believe are the source of most of the error people see. And do you leave the bubble or don't leave the bubble? Does the tip part count? By taking that little math shortcut, we have eliminated all of that error. Just so long as you do it the same way every time and with the same syringe, you are going to have accurate results. If you have an old test kit and use the 1.00mL titrant syringe to measure your 2mL (by filling to 1.00 twice) I'll let you keep all the way to the ones digit.

The precision is still a property of the test kit, and the titration types typically have a pretty high level of precision. But the accuracy is now a function of the standard solution you used. You can buy many different seawater standard solutions and often kits will come with a standard. You are at the mercy of the standard being right or not, but if you stick with one, then your tank should stay stable over a number of different kits.



HTH
 
Excellent David. :thumbsup:

All we need now is to find a place to buy a saltwater standard. Maybe Randy knows of another place that sells it. I checked that thread as well and apparently that site is down or out of business. ;)

This certainly will help out with all the test kit problems the hobbyists are having. :D
 
I found two companies by google and sent them e-mail to see if it's something that can be had.

osil.co.uk and seawatersolutions.com

We should also talk to Randy about which salts we would absolutely have to have to DIY it for test kit purposes. I would think we need to get at least sodium, chloride, sulfate, carbonate, borate, calcium, and magnesium. That might not be so hard to do.


Take this recipe:


23.98 g sodium chloride.........0.410mol
5.029 g magnesium chloride....0.0528mol
4.01 g sodium sulfate............0.0282mol
1.14 g calcium chloride..........0.0102mol
0.699 g potassium chloride.....0.0094mol
0.172 g sodium bicarbonate...0.0020mol
0.100 g potassium bromide....0.000840mol
0.0254 g boric acid..............0.000411mol
0.0143 g strontium chloride...0.0000902mol
0.0029 g sodium fluoride.......0.0000315mol
Water to 1 kg total weight.


For the purposes of making a standard for test kits, we could probably leave out NaF, SrCl2, KBr, and probably KCl.

We can substitute the boric acid with borax and acid. Really, since the concentration is so low, if it doesn't throw off the pH and alkalinity too much, we might get away with just a straight replacement with borax.

The only other thing on the list that we don't already have at our ready disposal is Na2SO4. We could instead replace part of the MgCl2 with MgSO4 and make up the sodium and chloride difference with NaCl.

That gives us a recipe of:

NaCl..........0.4664 mol..........27.2564g
MgCl2........0.0246 mol..........2.3422 g
MgSO4.......0.0282 mol..........3.3943g (6.9505g as heptahydrate)
CaCl2.........0.0102 mol..........1.1320g (1.4995g as dihydrate)
NaHCO3......0.0020 mol..........0.172g
Na2B4O7.....0.0001 mol..........0.0381g (as decahydrate)
HCl.............0.0002 mol..........2ml of 0.1M HCl

Dilute to 1kG total weight with water.

We should probably check with RHF about the borate and potassium, but something like this might be worth a try. We could tweak it a little if there's a problem.
 
Last edited:
This is the reply I got back this morning. It's not cheap.
Breaks down to $67 for 200mL, or about $0.67 to standardize a calcium or magnesium kit (2mL sample). Problem is you have to buy 10 bottles at a time. Perhaps people could go in together.

Lugene Curry lugene-curry@guildline.com to me
show details 8:52 AM (48 minutes ago)
Good Morning David

Your inquiry to OSIL re: seawater, has been forwarded to Guildline Instruments as we are the sole distributor of seawater in North America. Here is the pricing you requested:

P Series IAPSO Standard Seawater (item # 929-77777)
1 carton = 10 x 200mL glass bottles (must be purchased by the carton) $670.00

Quantity discount available
2 – 4 cartons $620.00 per box
5 – 9 cartons $590.00 per box
10 – 19 cartons $560.00 per box
20 – 99 cartons $540.00 per box
100+ cartons $530.00 per box

Delivery is approximately 10 days A.R.O. if stock. If not, delivery is approximately 3-4 weeks A.R.O.

Our terms are Net 30 days (with purchase order and 3 credit references for first time buyers) or we accept Visa and Mastercard. Shipping is collect on your courier/carrier account number or we can prepay and bill.

Thank you for your interest and feel free to contact me if you have any further questions or you wish to place an order.

Regards,


Lugene Curry
Guildline Instruments Ltd.
Customer Service Coordinator
21 Gilroy St., PO Box 99
Smiths Falls, ON
K7A 4S9
613-283-3000 ext. 100
613-283-6082 Fax
Toll-free 1-800-310-8104
Email lugene-curry@guildline.com
 
I'm curious as to how accurate the pure water standards would be for this. They may be close enough? They can be had for around $20.
 
I'm curious as to how accurate the pure water standards would be for this. They may be close enough? They can be had for around $20.

Even if they're not perfect, it would at least give us a standard to measure kits against and a way to compare two kits or two users. And as long as a person has a bottle, they can always make the new kit at least read like the old kit.
 
Sounds like the best option to me. ;)

Hach sells standards at what I believe are reasonable prices. Not sure if they have a mag standard or not. May be worth while to look into fresh water standards and let unfamiliar hobbyists know which are best for which test. Preferably ones that don't need to be diluted. :)
 
This piece from the attached PDF might bear some explanation.

232516 Calcium (NIST) as CaCO<sub>3</sub> (CaCl<sub>2</sub>)
vs.
2305442 Calcium (NIST) as Ca


So neither solution is made with calcium carbonate. They are both made with calcium chloride. The difference is the units.

For the first one, the units are being reported as equivalents of CaCO3, even though the solution is made with CaCl2. What this means is the amount of CaCO3 you would theoretically have to dissolve to get an equivalent calcium concentration. It doesn't make a lot of sense, but some of those old school guys get set in their ways and their units. So for the first one, you would convert to molarity by using the molecular weight of CaCO3 (even though it's made with CaCl2).

The second one reports calcium concentration as calcium, meaning to get it into molarity you use the molecular weight of calcium (40.078).

A few of the test kits use similar units for alk. Those that report alk as ppm are reporting it as ppm equivalent to CaCO3, even though in reality there's no way possible to ever dissolve that much CaCO3 into water.
 

Attachments

So that alkalinity standard may in fact be nothing more than 0.5N NaCO3, just reported as equivalents of CaCO3.
 
I have two different thoughts on the standard situation. First, I think that if there is any difference between the results in seawater and the results in one of the pure standards from Hach, then that would be good to know. That would give an indication of how much interference each kit is suffering, and that would tell us something about the quality of the chemistry being used.

But if we do find that there is significant interference, then it's probable that the manufacturer has that built into their calculations and we should too. In that case, we should try to use something that best simulates real sea water, at least in the ions that we expect to interfere. For instance, calcium and magnesium interfere with each others test.

When I get a chance, I think I'm going to make up some pure standards and some of that recipe a few posts up and try it out with my Salifert kits. If it works out, I'll go out on a limb and buy some of the Hach stuff to try out.
 
David,

Thanks for the input. ;)

It would be nice to see how the test kits respond to these standards. Perhaps it will provide a bit more insight and something to work with regarding all the test kit problems that seem to pop up almost daily. :)
 
If anyone of you are going to get a seawater standard sample count me in. David, if you are going to make some yourself, I can buy it from you. I can add another person to your standard pool on these test kits since I do also have Salifert kits. Although I am not a chemist by any means.

Alex
 
Oh by the way guys, great posts. Very informative. I learn the most from you guys including Boomer and Randy of course.

Alex
 
First round results

First round results

So I looked through the cabinet today to see what I can do. I don't quite have everything that I need, so I'll be ordering some things next week.

For starters, I don't have Mag Chloride other than what I have from BRS for my tank. The mag sulfate will have to come as a hydrate, so I am going to have to make up a solution and standardize that before I mix anything up for magnesium. The EDTA that I have is molecular biology grade, and a hydrate. That's hardly a primary standard. So I am going to order some more of that.

The calcium chloride I have on hand is reagent grade, but it is a dihydrate. So again not suitable. Depending on cost (I'll have to pay for these myself) I'll either buy some analytical grade anhydrous or more likely just make up a solution and standardize that against EDTA since I'm already having to do that for magnesium.

I'm not terribly concerned about the grade of the Sodium Bicarbonate. The manufacturing process for that stuff doesn't leave a lot of room for contamination. So I think I am going to trust what I have. I would probably be OK with Arm & Hammer, and we may be able to work out a DIY based on it.



So I decided today just to make up some solutions that may not be exact and just look at the feasibility of using pure standards as opposed to seawater standard. And I have to say, things are looking pretty good.



I made two solutions, for alk and calcium. I am using a balance that reads to 0.1mg and 100mL red volumetric flasks for dilution. The water is 18.2 Megaohm/cm water from our MilliQ unit and has been 0.2um filtered.


The alkalinity solution was made by adding 2.7512 g of Sodium Bicarbonate into 100mL of water followed by two 10:1 serial dilutions to yield a final solution with 0.275 g/L of sodium bicarbonate. This equates to (0.275 / 84.01) = 0.00327 M or 3.27 meq/L.


The calcium solution was made by dissolving 1.2932 g of Calcium Chloride dihydrate into 100mL of water followed by a 10:1 dilution to yield a solution with 1.293 g/L of Calcium Chloride (as dihydrate). This equates to (1.293 / 147.014) = 0.008795 M. Expressed as ppm Ca, this is (0.008795 * 40.078 * 1000) = 352.5 ppm. Since this was made with a known hydrate, and we have to assume that it actually has a little more or less water (probably more) than indicated, I fully expect this solution to read a little low. But it at least speaks to whether or not we can work this way with pure standards.





I brought these solutions home and ran them by my Salifert kits. Each test was performed 3 times and averaged. (They actually all got the same result. Easy average.)

The tests were run by the directions with the exception that the sample was delivered using a calibrated 1000uL Eppendorf pipette instead of the syringe they give you. This was done to take sample volume out of the equation. The titrant was delivered by the normal syringe provided with the kit.





For the Alkalinity solution, I expected 3.27meq/L and measured 3.19meq/L. That's not bad. Not bad at all for a test with an 11.5 meq/L level of precision and a solution that is standard only by my balance. I also checked the 6.7dkH standard solution that came with the kit (once) and did indeed get 6.7dkH.

So I think it's going to be safe to use a pure solution of bicarbonate to standardize an alkalinity kit. I'm interested to try the Hach standard.






For the Calcium solution, I expected 352.5ppm (I actually expected to read low since I made it with a hydrate) and actually measured 330ppm. Again, this isn't too bad given how I did this. I will say that the indicator color hue was slightly different from what it normally is when I test my tank. This may be due to the absence of magnesium. But the color change was still more than enough to find my endpoint. I also checked the 470ppm reference standard that came with an old SeaChem kit (one measurement) and got 460ppm.

I would like to run this one again once I get better reagents, but I think it's going to turn out that we can use the pure standards to work with the calcium kits too. I do reserve the right to check this solution against EDTA before I make a final decision.



I still need to look at magnesium, but I am going to wait until I have some mag chloride. So hopefully in the next couple of weeks. But I think things look really good for our being able to use the pure Hach standards to measure our test kits against. I'm interested to see how things work with other brands. I have some SeaChem kits that I should also give a try. I will keep you posted.
 
Thanks for taking the time to detail the math.

I love chemistry! And my reef. And your attention to detail.
 