New member
Hi Randy,
I just received one of these cheapy marine-depot refractometers:
http://www.marinedepot.com/md_viewItem.asp?idProduct=MD2101
It actually "seems" like a decent unit.
So, I want to calibrate it. The instructions just say to use distilled water to zero it out, but I've seen your comments in other threads in here about how this may not be the best way to calibrate it. Presumably, the idea here, is that the scale/readout may introduce some linear or curved error as a function of the salinity that's being measured, so you want to calibrate with a solution as close as possible to the target salinity you will be using the instrument to measure most of the time. This would cause the error to be ~0 AT the target salinity, with error increasing as the measured values stray away from the target...
This is my hypothesis, does this sound correct?
So, I'd like to do a couple of things. I'd like to first calibrate it with distilled (RO/DI with 0TDS) to zero it out. Then, I'd like to calibrate it using one of your calibration standards, and see if the calibration with the RO/DI water was off. This should show me the amount of error that my particular unit introduces when you're 0.0264 away from your target (or at least -0.0264 away from target).
For making the calibration solution, I'm a little confused about the difference between refractive index and spg. In your article, the refractometer solution table uses refractive index, but the scale on my refractometer is labled as sg. Here's a pic of it actually:
http://www.marinedepot.com/IMD/refractometer_md_logo_large.jpg
So, am I supposed to calibrate using the Refractive Index standard or the SPG standard from your article? I suppose I could just use the PSU scale?
Also, Do you know what that d20/20 symbol on my meter is?
One more thing. I assume these calibration formulas are linear, so I can multiply the amounts by a fixed constant. For instance, I don't have a scale, but I do have a 100mL graduated cylinder. So, for the salt, my best option is to use the teaspoon measurement of ~6.20g NaCl.
I bring this up, because the formula for the refractive index standard of 35PSU(1.3394 refractive index) calls for 3.65g NaCl and 96.35mL of water. The constant to get 6.20g (1tsp) of NaCl is 1.69863, which gives us 163.6mL. But, your formula for 35PSU(1.0266 spg) calls for 6.20g of NaCL and 161mL water... Acutally, now that I notice this, I see that in the %NaCl column in table 2 (refractive index) and table 5 (spg) from your article have different values for solutions of 35.0PSU. What am I missing here?
For reference, here's the article:
http://www.reefkeeping.com/issues/2004-06/rhf/index.php
Thanks so much for your patience and teachings,
Mathias
I just received one of these cheapy marine-depot refractometers:
http://www.marinedepot.com/md_viewItem.asp?idProduct=MD2101
It actually "seems" like a decent unit.
So, I want to calibrate it. The instructions just say to use distilled water to zero it out, but I've seen your comments in other threads in here about how this may not be the best way to calibrate it. Presumably, the idea here, is that the scale/readout may introduce some linear or curved error as a function of the salinity that's being measured, so you want to calibrate with a solution as close as possible to the target salinity you will be using the instrument to measure most of the time. This would cause the error to be ~0 AT the target salinity, with error increasing as the measured values stray away from the target...
This is my hypothesis, does this sound correct?
So, I'd like to do a couple of things. I'd like to first calibrate it with distilled (RO/DI with 0TDS) to zero it out. Then, I'd like to calibrate it using one of your calibration standards, and see if the calibration with the RO/DI water was off. This should show me the amount of error that my particular unit introduces when you're 0.0264 away from your target (or at least -0.0264 away from target).
For making the calibration solution, I'm a little confused about the difference between refractive index and spg. In your article, the refractometer solution table uses refractive index, but the scale on my refractometer is labled as sg. Here's a pic of it actually:
http://www.marinedepot.com/IMD/refractometer_md_logo_large.jpg
So, am I supposed to calibrate using the Refractive Index standard or the SPG standard from your article? I suppose I could just use the PSU scale?
Also, Do you know what that d20/20 symbol on my meter is?
One more thing. I assume these calibration formulas are linear, so I can multiply the amounts by a fixed constant. For instance, I don't have a scale, but I do have a 100mL graduated cylinder. So, for the salt, my best option is to use the teaspoon measurement of ~6.20g NaCl.
I bring this up, because the formula for the refractive index standard of 35PSU(1.3394 refractive index) calls for 3.65g NaCl and 96.35mL of water. The constant to get 6.20g (1tsp) of NaCl is 1.69863, which gives us 163.6mL. But, your formula for 35PSU(1.0266 spg) calls for 6.20g of NaCL and 161mL water... Acutally, now that I notice this, I see that in the %NaCl column in table 2 (refractive index) and table 5 (spg) from your article have different values for solutions of 35.0PSU. What am I missing here?
For reference, here's the article:
http://www.reefkeeping.com/issues/2004-06/rhf/index.php
Thanks so much for your patience and teachings,
Mathias