Ph Probe

Anyone? I have all the manuals including the new "Quick Start Manual" and if there are instructions in any of these about calibrating the Ph probe I can't find it. I know the probe has to be calibrated because the high and low calibrating fluid came with my Profilux3. Any help will be greatly appreciated.
 
Calibration for all probes are performed through the controller display.

1. Press any of the arrow keys on the ProfiLux
2. Select Sensor settings
3. Select the probe settings you wish to view
4. Select Calibration
5. Follow on-screen prompts
 