TDS meter calibration

I generally do not believe that TDS meter calibration is necessary or useful when used to monitor RO/DI. There are, in fact, several different ppm TDS unit scales, and few reefers actually distinguish them anyway.

The thing that gets off in a TDS meter is not the zero point, but just the slope, so if 100 ppm TDS really reads as 110 ppm TDS, then at 1 ppm TDS, the reading will be 1.1 ppm TDS (which rounds to 1 ppm TDS).

With that being the case, there are two things that you use TDS for:

1. Monitoring RO performance by the ratio of TDS in to TDS out. That ratio will be unchanged by calibration.

2. Monitoring when a DI is depleted. Since I recommend replacing the DI as soon as you detect a rise in TDS, it does not matter exactly whether the TDS is 1.0 ppm or 1.1 ppm or 2 ppm or whatever when you detect it: it is time to replace the DI when it goes above 0 ppm.

That said, there are a few spring water brands that put TDS readings on the label that may allow inexpensive calibration if you want to do so, but do not want to buy a commercial standard. So you might check in the grocery store next time you are there.
 
Well Randy used a lot more words than I planned to, but I don't ever calibrate my TDS meter, for the reasons he explains.
 