What Calibration solution do you use for your Refractometer?

Your refractometer may be off as much as 3-5 points calibrating it with distilled water. You want to use the Pinpoint 53.0 mS solution and calibrate to 35 PPT or 1.0264
 
I thinkg most people use RO/DI water to calibrate, and alot of the instructions that come with them state to do it that way, but to be more precise, any of the above solutions work well. I use the Pinpoint 53.0
 
<<< I thinkg most people use RO/DI water to calibrate, and alot of the instructions that come with them state to do it that way, but to be more precise, any of the above solutions work well. >>>


Actually the instructions are incorrect if you're using your refracto for seawater testing. Using RO/DI or distilled to calibrate it to 0 and you may end up off as much as 3-5 points because the refractos are made for testing brine (salt) solutions and must be calibrated differently when testing seawater, hence always use the 53.0 fluid and calibrate to 35 PPT :-)

There are of course a few refractos that will actually be close when calibrated with RO/DI or distilled, but why take the chance of being way off?
 