I spent months battling with incorrect conductivity readings until I followed this procedure. Now I'm bang on the money
Try not using temp comp. If those steps give you equal results try the steps here. You an use the same calibration solution for a week or so. No need to open new bags just to do this each time.
1. Remove the Cond Probe from your PM2.
2. Remove the temp probe from your PM2. (If ANY)
3. Then go through the calibration again.
4. Aquacontroller Display : SETUP> COND SETUP
5. Set the Temp Comp to 0.0.
6. Click on COND CALIBRATE
7. Fail purposefully. It will reset to Defaults by clicking on SAVE, OK, SAVE, Etc.
8. Now Plug the Cond probe into your PM2
9. Now go to SETUP> COND SETUP> COND RANGE > SALINITY
Make sure the cond probe is plugged all the way in. Go to SETUP> COND SETUP> COND CALIBRATE.
1. It will say 'Dry Settling' This is the Dry Phase. Hold the cond probe out in the air for a minute and wait 1 minute. Then click OK
2. Now select your solution. use the up and down arrows to select 53000 then click OK
3. Then place the cond probe in the bag of solution. Keep it there for about 2 minutes then press OK.
4. It'll save and bring you back out to the menu. Click HOME and you're all done.
5. Keep the probe in the bag of solution and what does the cond probe read? It should read 34.9 or 35 ish.
If it read 34.9-35.0 in the solution the probe is capable of reading accurately and is operating correctly!