Anyone here ever configure new ORP probe on Archon or have documentation on how to calibrate new ORP probes?
Does this still hold true:
ORP/pH "“ These inputs can be individually programmed to be either ORP or pH. When the Dip switch is set to the "œup" position, it is set for ORP. "œDown" sets the
port for pH.
I have read this http://digitalaquatics.com/downloads/Archon/Archon_Rev3.pdf but I remember the last time I got a orp probe 2 years ago DA support said Dip switch changes was not required anymore after a firmware update.
Instructions:
Change the channel name to "œORP" or your desired name, then click the "œSet port to ORP mode" box and click Save. Wait for the page to refresh (about 5 seconds).
ORP Calibration
Click the Calibration Start button
Confirm that you see only one set point, and that the target is 400, and that the Set port to ORP Mode box is checked
Soak the ORP probe in the calibration solution for 20-30 minutes, then click the Copy button next to the "œ400" Target (this will copy the raw data numbers
into the Data box for you).
Click Finish to write the data to the Archon unit. Calibration is complete.
Does this still hold true:
ORP/pH "“ These inputs can be individually programmed to be either ORP or pH. When the Dip switch is set to the "œup" position, it is set for ORP. "œDown" sets the
port for pH.
I have read this http://digitalaquatics.com/downloads/Archon/Archon_Rev3.pdf but I remember the last time I got a orp probe 2 years ago DA support said Dip switch changes was not required anymore after a firmware update.
Instructions:
Change the channel name to "œORP" or your desired name, then click the "œSet port to ORP mode" box and click Save. Wait for the page to refresh (about 5 seconds).
ORP Calibration
Click the Calibration Start button
Confirm that you see only one set point, and that the target is 400, and that the Set port to ORP Mode box is checked
Soak the ORP probe in the calibration solution for 20-30 minutes, then click the Copy button next to the "œ400" Target (this will copy the raw data numbers
into the Data box for you).
Click Finish to write the data to the Archon unit. Calibration is complete.