Is it possible to make my own TDS calibration solution?

weblance

New member
I have a HM Digital TDS meter: http://www.tdsmeter.com/products/tds3.html and it needs to be calibrated. Since I live in Denmark the right calibration solution (http://www.tdsmeter.com/products/calibrationsolution.html) is very hard to get.

Is it possible for me to make my own solution? I'm thinking of using this formula:

1 mg NaCl/liter of distilled water = 1 ppm TDS

Am I right or can't this be used? If I'm right - where can I get NaCl, is it just regular table salt or can I get it at a pharmacy?

Thanks in advance!

/Carl
 
In general, if you use the TDS meter to monitor RO/DI devices, it isn't normally necessary to calibrate.

Yes, IMO you could use table salt to make a standard. The two ways would be with a good scale and small amount of table salt in DI water, or a larger amount measured by teaspoon that you lowered in TDS by serial dilution. I have not worked out details of the latter however.

What range of TDS does your meter cover?
 
I think that it needs to be calibrated. It reads 1 ppm on the RO water, and 1 ppm is pretty low even for a RO unit.
 
Yes, if you can measure out that amount, that should be an adequate way to calibrate the meter. :)

The weight measurement error will be smaller if you use a larger amount, and calibrate at, say 5,000 ppm (= 5 g/L).
 
No, it is not enough to be significant, although you can get table salt without it. The other ingredients also should not be a problem, since we are not expecting or needing to be closer than +/- 10% (or more). :)
 
<a href=showthread.php?s=&postid=6769589#post6769589 target=_blank>Originally posted</a> by Randy Holmes-Farley
Yes, if you can measure out that amount, that should be an adequate way to calibrate the meter. :)

The weight measurement error will be smaller if you use a larger amount, and calibrate at, say 5,000 ppm (= 5 g/L).

I am interested in this also. I have the same meter. I wanted to verify mine as I had a defective RO membrane. It is next to impossible to find a calibration soultion that is 342 ppm.

How much is 5 grams of say Morton Salt in teaspons?
Is store bought distilled water actually "0"ppm?

As of today my ro/di is reading 0 ppm so I could use that and how ever much salt in a liter??
 
No, I do not think so. That's just what the web site above says their calibration solution is. I last used a packet conductivity solution of 10 micro something:)
My tap water is about 342ppm.

Now that I think about it, if my meter is reading "0" and calibrated to the 10. Can I just mix a trace amount of salt in a bottle of RO/DI and whatever ppm that comes to, use that as a check or standard? It shouldn't change over time if sealed will it?

If we estimate the amount of Morton of salt to add to a liter of RO/DI to get somewhere close say 300 to 500 ppm that would be
at least be in the ball park.
 
As long as your believe it is calibrated, then yes, it is trivial to make your own standard from DI water and some table salt. Good point! :)
 