Probe calibration problem

gil716

New member
I just spent an hour trying to calibrate 3 different ph probes on my ACII and I'm out of ideas. During the "settling phase" for 7.0, it measures in the 6.x range - not unusual. I hit select, place the probe in the 10.0 solution, hit select again and for some reason the probe still settles in the 6.x range.

Again I tried this with three different probes. One of which calibrates correctly on a pinpoint monitor. Anyone else ever experience this?
 
Finally able to do a search at 2 in the morning. Someone mentioned that the temp probe should be in the solution while calibrating, but another post said that this doesn't apply to the ACII. The quest continues.
 
well, looks like that was the issue. recalibrated with the temp probe in the solution and everything calibrated correctly.
 