Is that true of the other probes as well? I'm getting really spiky readings from my pH and temperature probes. I get 2-3 bizzare spikes per day. They are both in an area that have a few bubbles...
Temp correction along with calibration? I called Carl and he said yes use temp correction. Set it for 2.1 then calibrate. I moved probe away from pumps in the sump. Also away from other probes. I used Neptune 53000 calibration fluid. Got the fluid to 25c. Have the temp probe in the same space as the COND probe connected to PM2. calibrated w Carl on the phone everything seemed to work. After calibration my salinity was high. Almost 37ppt. So before I panic I measured salinity with my new DD Refractometer (ATC) it's a temp correction seawater refractometer. Calibrated with RO/DI water @ 20c. The refractometer is at 20c as per instructions. The refractometer is reading about 2.1 lower than the probe. Probe is 36.9 and refractometer just under 35ppt. What is going on? Should I recalibrate and forget temp correction. As far as I can tell the temp never fluctuates more than .5 - 1F.I haven't seen any data 1 way or the other on which probe is more or less sensitive to electrical interference and I don't run a grounding probe but I wouldn't expect that to cause any issues 1 way or another in most cases. if this was an AC3 I might think otherwise but not with an Apex.
After calibration my salinity was high. Almost 37ppt. So before I panic I measured salinity with my new DD Refractometer (ATC) it's a temp correction seawater refractometer. Calibrated with RO/DI water @ 20c. The refractometer is at 20c as per instructions. The refractometer is reading about 2.1 lower than the probe. Probe is 36.9 and refractometer just under 35ppt. What is going on? Should I recalibrate and forget temp correction. As far as I can tell the temp never fluctuates more than .5 - 1F.
Well the DD seawater ATC refractometer states RO/DI water should be used and they have a reference point on the scale for it ( Zero @ 20C) . I'm so frustrated that I just bought 35ppt solution. Funny I have some Pinpoint 53000 and that seems to be off as well. But someone said that's for probes not refractometers. But that stuff didn't work on the Neptune probe either unless my watr is like 1.012! Killing me!Calibrating your refractometer with RO is not the most accurate method. Many refractometers are quite off around 35ppt when using this method of calibration. To be the most accurate, get yourself some 35ppt calibration solution for your refractometer such as this : http://cgi.ebay.com.sg/35-PPT-REFRACTOMETER-CALIBRATION-SOLUTION-/270370466481
My bet is on the Salinity probe reading being more accurate...![]()
OK if thats the case I assume the ATC refractometer should be calibrated at about 68C. The reference fluid temp shouldn't matter. If thats the case my refratometer is calibrated correctly and DD is correct RO/DI water for this unit works. I'm dead on. So now my tank is 35ppt dead on w refractometer and Neptune is a 37.3, TEMP COMP 2.1. I'm going to recalibrate with TEMP probe in the same water as COND probe which has the 53000 solution floating in it overnight. I will leave the the TEMP CORRECTION at 2.1 and see what flies.The Pinpoint 53000 solution can be used to calibrate your refractometer.
You cannot use the Neptune probe calibration solution to calibrate the refractometer however...![]()
Geez your refractometer sounds like a pain in the butt. I've got an el-cheapo one that I haven't calibrated in a year and a half. I don't bother with worrying about temp with it and it reads within .1-.2 of my classic glass hydrometer, a swing-arm hydrometer and my salinity probe.OK if thats the case I assume the ATC refractometer should be calibrated at about 68C. The reference fluid temp shouldn't matter. If thats the case my refratometer is calibrated correctly and DD is correct RO/DI water for this unit works. I'm dead on. So now my tank is 35ppt dead on w refractometer and Neptune is a 37.3, TEMP COMP 2.1. I'm going to recalibrate with TEMP probe in the same water as COND probe which has the 53000 solution floating in it overnight. I will leave the the TEMP CORRECTION at 2.1 and see what flies.
It's unreal to me that 500.00 worth of lab grade probes and controllers and it can't measure salinity. Frankly disgusted. An 89.00 refractometer is more accurate huh? I'm so annoyed. Should have used the AI controller and the refractometer and saved 400.00. It seems you went through the exact same thing I'm going through right now. Except I won't send the rmeter back now. There is exactly 2.1 difference in readings which is the 2.1 temp correction in the probe. But of course it should have nothing to do with it. I am going to try and calibrate probe w/o temp correction for kicks and see what happens. I'm obsessed with the 2 matching.... LolTrust your refractometer, not the salinity probe.
When I started up my first saltwater tank (about 3 years ago), I went through absolute hell to figure out my salinity. I was using the AC3 Pro w/ the Lab Conductivity probe and I couldn't get a consistant reading.
I took water samples to every LFS in town and oddly enough the variance in calibration of refractometers was significant. I got readings from 1.024 to 1.027.
I bought every brand of calibration solution available. I calibrated my refractometer with the pure water that came with it, RO/DI from my system and 35ppt solution. I calibrated my salinity probe with solutions from multiple manufacturers.
No matter how hard I tried, I could not get the salinity probe to agree with my refractometer, so I sent my refractometer back and had it replaced under warranty. I had exactly the same problem with the new one.
This time I took my refractometer to every LFS in town and compared my measurements of their tank water to my refractometer. I went to the best, most reliable LFS last and my calibration was dead on with their refractometer.
Considering this LFS is the source for most of my fish and coral, I accepted this reading as valid and have used it for the baseline ever since.
I never did get the salinity probe to agree with the refractometer.
Moral of the story - Use the salinity probe to detect change, not as the baseline. If you want, you can always calibrate it using your tank water as the baseline.
I assume you're throwing out the $500 for dramatic effect. I'm also assuming you bought an Apex for more than JUST salinity measurement. Although between the PM2 and the probe it is over $200. So yeah it would be nice if it was more accurate, but that's just how those probes work. The conductivity probe I had for "that other controller" behaved the same, perhaps even a little worse. It's either that or no option for salinity whatsoever. It's the only technology available for Neptune to make a constant salinity measurement feature. Although mine behaves itself fairly well most of the time, so I'd like to think you can get yours working more reliably somehow.It's unreal to me that 500.00 worth of lab grade probes and controllers and it can't measure salinity. Frankly disgusted. An 89.00 refractometer is more accurate huh? I'm so annoyed. Should have used the AI controller and the refractometer and saved 400.00. It seems you went through the exact same thing I'm going through right now. Except I won't send the rmeter back now. There is exactly 2.1 difference in readings which is the 2.1 temp correction in the probe. But of course it should have nothing to do with it. I am going to try and calibrate probe w/o temp correction for kicks and see what happens. I'm obsessed with the 2 matching.... Lol