This post got me to search some on the Hanna meter and someone is suggesting that the manufacture is claiming it has a +/- 0.04 error. Have you seen anything about that?
The accuracy of the Hanna meter is +/-0.04mg/L +/-4% of reading.
This is Hanna Specifications.
No one I have talked to can agree whether this means +/-0.04 of the final reading or does it mean +/-4% of the final reading. These are two completely different accuracies.
For example if my final test results with the Hanna were:
.10 and I had an accuracy of +/-0.04 then I would be in this range
( .06 - .14 )
But if I had a final test result with the hanna of:
.10 and I had an accuracy of 4% of reading then I would be in this range
( .096 - .104 )
I have never sent an email to hanna for an explanation and I havent read of anyone getting an explanation from them.
So this is what I did/do to test my hanna.
I test distilled water which is zero PO4. This should test 0.00 on my Hanna. I also test my RO/DI water straight from the DI line of my unit. This should also read 0.00
With that said I have never gotten my hanna to read 0.00
It has always read .01 when testing both these samples.
This is plenty accurate for me. And it is possible that both the distilled and my RO/DI has a little bit of PO4. THe distilled water could have a little bit of PO4 leached out from the plastic jug it is stored in and my RO/DI is probably not perfect.
I read all the time about how the hanna is not accurate and I just dont agree. For my purposes it is absolutely a great tool for measuring phosphates.
It will outperform any and I mean any hobby test kit I have used. And I have used a few. I recommend the Hanna HI 93713 Phosphate Low Range ISM.