Adjusting PH Calibration Data?

zhangg3

Member
After I upgrade firmware to 5.0.6 I then loaded settings from a previously saved settings file to PIII. However it did not load the PH Calibration values. I had copied those numbers down but the ADC1 and ADC2 boxes are not editable.

Is this something that can be done?
 
you can, save your sensor data to your desktop, now open that in notepad.

look for:
12, "CODE_SENSORPARA0ITEM2", 452
13, "CODE_SENSORPARA0ITEM3", 250

change the 452, 250 with your values...
PH2 on a EX box is at:

1012, "CODE_SENSORPARA8ITEM2", 422
1013, "CODE_SENSORPARA8ITEM3", 227

save and re-upload
 
Saving and reloading edited sensor data works. Although I ended up doing calibration since I decided to do some maintenance anyway.

Thanks for your help!
 
yes unless you are testing different firmware, most time the calibration is required anyway for proper reporting :)
 