Difference in PH reading between PH1 & PH2

blk94rs

New member
Im setting my AC3 up today...I calibrated both of my brand new Neptune standard grade ph probes exactly the same way. I waited for the numbers to settle in both the 7.0 & 10.0 solution for a full minute (without any change) before pressing the 'select' button.

I placed both probes in my sump after calibration to see if the were close. PH1 reads 8.04 & PH2 reads 8.17. Is this anything to be worried about?

I will be using PH2 to control the co2 solenoid for my calcium reactor.
 
When you calibrated them did you calibrate each probe, connected to the port it will be used on? I've seen where people calibrate a probe on (say) PH1 connection, then disconnect it and calibrate the 2nd probe on the same connection; that would cause a problem as calibration offsets are per connection and probes cannot be changed after calibration.
 