I'll be interested in seeing which model they might have.
For those interested, I did receive a technical recommendation from a gentleman at Cole-Parmer.
He recommended that I use an instrument which is calibrated at 60'F and then apply a correction factor using the following table:
From this chart, the relevant portion for my Reef Tank is this:
It states that interpolation is required for temperatures and SG readings not specifically listed in the chart.
I did some calculations and found that it's not a linear formula and there is a sharper curve the higher the temperature becomes.
However, when I plugged in the SG correction factors for the temperatures of 20'C and 30'C (which skips 1 charted interval), and then solved for the mid-point of 25'C, the results were very accurate so within this Reef Tank range I am happy to interpolate as if it was linear.
This iPhone app is called Interpolator and was easy to use.
The correction factors are within a narrow range and don't change much at all between an SG reading of 1.000 (RO/DI water) and 1.100 (Salty soup with corals not very happy).
At 27'C and 1.000 SG, the factor is 0.0022
At 27'C and 1.100 SG, the factor is 0.0023
So the SG isn't really a factor, you don't even have to worry about moving up or down on the chart depending on what you read since it's only a 0.0001 difference.
The temperature is what makes the difference and needs to be carefully checked prior to testing with a NIST calibrated thermometer I would think. Or at least one you have checked in boiling water and ice water to see how far off it is.
You can see below that if I use 1.024 SG as the reading I'm getting on the Hydrometer (quite likely to be the case), the correction factors change by 0.0005 depending upon a temperature swing of 2'C.
At 27'C and 1.024 SG, the correction factor is the familiar 0.0023
However at 25'C and the same 1.024 SG, the factor drops down to 0.0018.
I will simply use 0.002 as my factor when the temperature is between 26'C and 27'C.
Since my tank ranges between 26.5 to 27.5, this seems like a reasonable temperature that the water will be at the time of testing in the glass tube.
So I will be looking for a reading of 1.024 SG on the Hydrometer, to tell me I'm at 1.026 in my tank.
Cole-Parmer recommended the following Instruments for my application:
I would agree with this since our target measurement range is in the middle of the 1.000-1.050 scale, and it appears to have 0.0005 graduation lines which is great.
I also asked them for a suitable test tube and was recommended this product:
I think I will place the order and see how it differs from the cheaper Amazon.ca Hydrometer and Test Tube that I already ordered while awaiting this reply. The Amazon Hydrometer was $2.50 and the test tube was $30.
But I would like to have one in my water mixing room as well so hopefully it will be suitable.
I'm assuming the lab-grade stuff shown above will be top notch quality, for those prices.
Now for the big question... do I need a secondary correction factor to account for all the "stuff" in Seawater which causes refractometers to read Seawater differently than Brine? Or is that only an issue with refractive index and not a problem when reading SG with a floating Hydrometer?