Kroghs

Here it is"
I would like my doser to go on at 3pm daily and dose for 6 minutes
Fallback OFF
OSC 900:00/006:00/904:00 Then on

Also it seems to be taking quite some time to calibrate my conductivity probe, numbers still changing, currently at 0543 and slowly going up, is this normal. It has been in the calibration solution for about 15 minutes.
 
I would like my doser to go on at 3pm daily and dose for 6 minutes
Fallback OFF
OSC 900:00/006:00/904:00 Then on
To dose once per day using OSC, the sum of all three time parameters needs to add up to exactly 24 hours (1440) minutes. Currently, your total OSC cycle is 1810 minutes long, or a bit over 30 hours.

This is what you need:

OSC 900:00/006:00/534:00 Then ON
 
Yaaa, I finally understand, thank you very much! Calibration still going up, is that okay? Also my solenoid keeps tripping my gfi outlet. Just installed the gfi outlet with the Apex, first step, replace gfi? Might be bad or I damaged it during installation? You agree?
 
Just wait until the calibration is steady. Also when calibrating, the calibration fluid must be the same temperature as the water that the temp probe is in. I usually float my cal fluid in my sump for a while to bring it to the same temp before calibrating.

As for the GFCI I would suspect that the solenoid is faulty. Test the solenoid on another GFCI. Possibly you have another GFCI installed somewhere in your home like in a bathroom. If it trips that one too then you know that the solenoid is faulty for sure.
 
Thank you for all your help, may I ask for more help as I continue setting up my Apex system? Next challenges will be powerheads, lights and pumps?
 