Yes you fixed it
The Sigma is only for doing SG at auto correct and non-autocorrect.
Ok now
25.3 sigma-15.
It is pretending to correct to a SG meter that is = to a calibrated point at that temp of 18.90C or 25.00C or what ever the sample temp is. Where you want it
Sigma-t
24.3 sigma-t and 22.8 sigma-t. It is in a non temp correct mode for SG. It is telling your or pretending the calibration temp point is 15 C and is not adjusting for 18.90C or 25.00C . You do not want it in this mode
It is currently set to 1.90%/C (at the factory
Yes leave it. This is in some meters to correct that pre-programmed cure I talked about.
Should I change it seeing that my temperature compensated conductivity value is not 52mS/cm when not on 25C?
No it is fixed. Leave it as is. All your readings now are correct, like perfect. No more playing. Put back in box
@ 25.00C in sigma-15 mode
==================
52.01mS/cm (Temp Comp. Cond.)
52.02mS/cm (Actual Cond.)
34.22 Salinity
25.4 sigma-15
We are fixed I think you did good :thumbsup:
Ok now
25.3 sigma-15.
It is pretending to correct to a SG meter that is = to a calibrated point at that temp of 18.90C or 25.00C or what ever the sample temp is. Where you want it
Sigma-t
24.3 sigma-t and 22.8 sigma-t. It is in a non temp correct mode for SG. It is telling your or pretending the calibration temp point is 15 C and is not adjusting for 18.90C or 25.00C . You do not want it in this mode
It is currently set to 1.90%/C (at the factory
Yes leave it. This is in some meters to correct that pre-programmed cure I talked about.
Should I change it seeing that my temperature compensated conductivity value is not 52mS/cm when not on 25C?
No it is fixed. Leave it as is. All your readings now are correct, like perfect. No more playing. Put back in box
@ 25.00C in sigma-15 mode
==================
52.01mS/cm (Temp Comp. Cond.)
52.02mS/cm (Actual Cond.)
34.22 Salinity
25.4 sigma-15
We are fixed I think you did good :thumbsup: