Ted_C
Active member
Hi All.
I seem to be having some issues calibrating my salinity probe. I've repeated this process three times now and have been getting the same result with 3 separate 53 ms solutions.
I do have the PM2 module w/ a salinity probe that is approx 2 years old and a temperature probe.
The probe was dry for about a year - as I took a break from the hobby. I have attempted the calibration when first wet (maybe a day of running in water), after a week running in water, and two weeks running in water.
The solutions I've been testing are:
1 foil pack of the 53 ms Neptune solution
1 Pinpoint 53 ms Solution that measures to 35 PPT on the VeeGee Refractometer
Another pinpoint 53 ms solution that also measures to 35 PPT on the VeeGee
My settings when calibrating are
Salinity
Temp Comp: 2.0
When I calibrate with the 53 ms solutions - the same thing happens:
I rinse the probe with RODI water
Let it dry
Calibrate to zero - the probe reads -0016 after it stablizes
Calibrate to the 53000 solution
The probe reads 0625
After accepting the value and examining the main screen - my salinity reads around 9.2 - while it is still in the 53ms standard solution.
Any ideas to try or is the probe junk at this point?
I seem to be having some issues calibrating my salinity probe. I've repeated this process three times now and have been getting the same result with 3 separate 53 ms solutions.
I do have the PM2 module w/ a salinity probe that is approx 2 years old and a temperature probe.
The probe was dry for about a year - as I took a break from the hobby. I have attempted the calibration when first wet (maybe a day of running in water), after a week running in water, and two weeks running in water.
The solutions I've been testing are:
1 foil pack of the 53 ms Neptune solution
1 Pinpoint 53 ms Solution that measures to 35 PPT on the VeeGee Refractometer
Another pinpoint 53 ms solution that also measures to 35 PPT on the VeeGee
My settings when calibrating are
Salinity
Temp Comp: 2.0
When I calibrate with the 53 ms solutions - the same thing happens:
I rinse the probe with RODI water
Let it dry
Calibrate to zero - the probe reads -0016 after it stablizes
Calibrate to the 53000 solution
The probe reads 0625
After accepting the value and examining the main screen - my salinity reads around 9.2 - while it is still in the 53ms standard solution.
Any ideas to try or is the probe junk at this point?