This threa has just made me check something that I'd been wondering about for a while. Why (I wondered) do the instructions tell you to calibrate with RO at 20C? Surely if it has ATC then you can calibrate with RO at any temperature within the ATC range and the job will be a good 'un?
It seems not! Based on what I read on the Aquarium Solutions website here
http://www.theaquariumsolution.com/d-d-refractometer-calibration, it seems that the ATC will sort out a salinity reading in the 'normal' range for any temperature, but the zero will appear off for any temperature other than 20C.
I hadn't realised this - I assumed that the ATC would pull all points (including 0ppt) onto the correct line at all temperatures. Now I know better!
I wonder if this is why some people get perfect readings with this refracto when calibrating with RO and others get it 1.5ppt out? When I got a perfect reading when calibrating with RO and then testing with a reference solution, perhaps I was just 'lucky' that I had miscalibrated with RO a couple of degrees off 20C and this made things line up perfectly. I fear I detect another round of calibration testing coming up here!
Or perhaps I'll just cal with a 35ppt reference and be done with it! :deadhorse:
Peter