Hi everyone,
I have been using an industrial pH controller/meter on my tank for the past 5+ years, and have had no reason to doubt it's readings - I replace the electrode every 18 months, and calibrate the unit monthly with pH 7.01 and 10.01 buffers from Hanna Instruments.
I have recently purchased a Hanna Instruments Mod. C99 bench Photometer, mainly to test for phosphate, but would like to also make use of some of it's other capabilities, such as testing pH (as a check on the electronic meter).
After a fresh pH electrode calibration, my tank water was reading 8.1 on the digital (electronic) controller. The photometer, using the Hanna HI93710-01 Phenol Red reagent, gave an "over-range" error, thus indicating that the tank water was at or above a pH of 8.5
Now, a discrepancy of +0.4 pH units is quite large, and I would like to know the following:
- Which instrument is most likely to be accurate?
- If the fault is with the photometer, would this be caused by something in the salt water interfering with the test? if so, what?
- is there something I can do to check the accuracy of the photometer - perhaps make up a solution of a known pH within it's operational range of 6.5 - 8.4, and test both instruments against this (I am guessing that using one of the calibration buffers would not be a good idea, as being highly buffered and already containing a color indicator, it would not work with the phenol red...)
Any further advice on using the photometer will also be appreciated - perhaps I'm just being stupid and not doing the test correctly (although I am following the instruction manual to the "T")
Thanks for any responses.
Hennie
I have been using an industrial pH controller/meter on my tank for the past 5+ years, and have had no reason to doubt it's readings - I replace the electrode every 18 months, and calibrate the unit monthly with pH 7.01 and 10.01 buffers from Hanna Instruments.
I have recently purchased a Hanna Instruments Mod. C99 bench Photometer, mainly to test for phosphate, but would like to also make use of some of it's other capabilities, such as testing pH (as a check on the electronic meter).
After a fresh pH electrode calibration, my tank water was reading 8.1 on the digital (electronic) controller. The photometer, using the Hanna HI93710-01 Phenol Red reagent, gave an "over-range" error, thus indicating that the tank water was at or above a pH of 8.5
Now, a discrepancy of +0.4 pH units is quite large, and I would like to know the following:
- Which instrument is most likely to be accurate?
- If the fault is with the photometer, would this be caused by something in the salt water interfering with the test? if so, what?
- is there something I can do to check the accuracy of the photometer - perhaps make up a solution of a known pH within it's operational range of 6.5 - 8.4, and test both instruments against this (I am guessing that using one of the calibration buffers would not be a good idea, as being highly buffered and already containing a color indicator, it would not work with the phenol red...)
Any further advice on using the photometer will also be appreciated - perhaps I'm just being stupid and not doing the test correctly (although I am following the instruction manual to the "T")
Thanks for any responses.
Hennie