the waste/product ratio will effect the TDS of the product water coming from your RO membrane. In a standard unit you should be running 4-5x as much water out of your waste line as is coming out of the product line. Like my example above, I'll explain more fully this time. My system water is about 125-150 TDS. When I first hooked up my most recent RO membrane the waste/product ratio was 2/1 and the resulting TDS of the product water was 8. When I properly adjusted the waste/product ratio to 4/1 it dropped the TDS of my product water to between 1 and 2. Basically, restricting the product line slightly allows more free flow of water through the waste line and keeps small particles from being forced through the RO membrane.
I'm pretty sure the standard rejection rate for RO membranes is around 98-99%. So, if your input water is 450 TDS and the product 24, that's still only 94-95%. So, that tells me you need to restrict your product water line more to get the rejection rate higher, so you see maybe 5-10 TDS product water. This will make your DI resin last longer. And the reason I keep saying to check the meter is because DI resin is extremely effective. If everything is hooked up correctly, and the resin is new, it should drop that 24 TDS water to 0 with no trouble, although this isn't good because it will deplete the resin faster than normal, which is why I said you should probably adjust the waste/product ratio. Adjusting the waste/product ratio isn't going to fix your problem, but it looks like a secondary problem that should also be fixed.
Bottom line though, if you're reading 24TDS into DI and 24TDS out of DI, there's only 3 possibilities. 1. Bad DI resin 2. Bad/wrong connections 3. Bad TDS meter