Osmolator 3155 - Problem, Red led lit, Audible Sound On

ronone10

Premium Member
Bought a used osmolator and it was tested before I paid for it. I decided to give it a bath, clean the parts. Now, when I plug in the controller, the red led light 'too high' is on and so is the warning sound.

According to the manual, it could be:

a) too much water - nope, running it dry
b) water level increased because you inserted your hand - no again
c) float ring of the sensor was turned 180 degrees during last cleaning. - not sure what this means, I didn't touch the float sensor, only the operating sensor. I pulled it out of the plastic holder and cleaned it.
d) safety sensor or controller is defective - I hope it isn't the controller. I don't know how to check the safety sensor.

Any suggestions on what is wrong?
 
red light on means your float sensor has been tripped. Make sure it's not stuck or something. Choice C is incase someone put the float sensor back on upside down.
 
There is a C clip at the top of the post the float is on, it can be removed and rotated 180 degrees. The other possibility is there is a magnetic field near the float switch or controller.
 
Thanks for the advice. After leaving it unplugged for 24 hours, I plugged it in and the audible sound stopped. It is working now. Strange, not sure what happened but the next time I have an electronic issue, I plan to unplug the device for at least 24 hours... :)
 
My guess would be it will recur, probably a ballast or something nearby is emitting a magnetic field, watch for the alarm coinciding with a set of lights or some other device turning on.
 