how can a probe that can't be calibrated be showing shifting numbers? can someone please explain this to me?
The calibration process measures the probe output at high and low points and expects there to be a minimum range between the 2 points; it is the readings - difference that determines the accuracy; yes, I'm over simplifying! You can have a probe that reads 'something' and still will not calibrate if the high and low calibration values are closer than tolerance woud allow; in other words the values read but the difference is so small between the points that the readings are inaccurate.
This all said; 1 bad calibration may not mean the probe is bad; it can also be that the calibration solution is bad. For example; if you calibrated with 2 solutions at the same value the calibration would fail so if the fluid was bad that would give a similar indication; the calibration would also fail. Since you mentioned you have 2 probes try 2 and verify it calibrates (that will somewhat test the fluid), then soak the other probe in vinegar for 1/2 hour, rinse blow off (dry) and try calibration. If it still doesn't work I'd then say you have a bad probe.