In June, I bought a Neptune 'lab grade' Ph probe to hook up to my Pinpoint Ph meter. It seems to need calibrated too often. The latest calibration lasted only 2 weeks. After calibration (I used Atlas calibration solution), 3.8 grams of borax in a liter of purified water showed Ph 9.28 instead of 9.18. Temp was very close to 77 F. http://web.archive.org/web/20021015005420/www.animalnetwork.com/fish2/aqfm/2000/feb/bio/default.asp
I double rinse the probe in RODI water between calibration solutions and am using a battery for the meter because I had interference when I used a wall wart.
Do you agree the probe is probably junk?
If so, can you recommend a decent probe?