Is it possible to change the calibration solution variable? The current options are 220mV to 250mV. I have some AM Pinpoint ORP calibration fluid, but it is 400mV, outside of the ProfiLux range. Is it possible to adjust the calibration range so that I can use the Pinpoint fluid? Is there an advantage to using a lower mV calibration fluid over a higher mV fluid.
I got the Redox probe up and running last night. After calibration, the controller measured 3mV with the null attachment and 216mV in the GHL 220mV calibration solution. These variances seemed reasonable to me.
However, when I tested the probe in the 400mV Pinpoint solution, the probe read 360mV, which seems like a fairly large variance. Is there an explanation for this amount of variance? Is it reasonable?