BeanAnimal said:Habib,
Thank you for taking the time to try and explain the deviation. ........
...........
Thanks!
leebca said:Habib,
What would explain the Seachem standard giving their test kit the same results; the LaMotte Test kit the same results; but the Salifert kit 25% higher results?
Should not a standard, regardless of 'systemic' issues give the expected results? Many of the examples you give is measuring the unknowns and measuring against Salifer internal standards. What do the Salifer internal standards read on the LaMotte and Seachem test kits?
I still have a concern when my LaMotte Test Kit reports X Ca; my Seachem TK reports X Ca; and the Salifert TK reports X + .25X Ca. The color change with the Seachem TK is very clear. Enough so that I can tip-off partial drops and see a change.
Thank you. Hope you had a good vacation!![]()
Perhaps the Seachem is adjusted to their reference or the reference value is adjusted to their kit or perhaps another scenario? Hard to say.
I still have a concern when my LaMotte Test Kit reports X Ca; my Seachem TK reports X Ca;
I don't have experience with the Lamotte (and also not with the current Seachem version). However, I explained in my previous post that the Lamotte requires a 12.9 fold dilution with DI water. A slightly premature end-point by 8 ppm will give a 12.9 x 8 = 103 ppm too low value.
When we use various standards and also various natural seawater samples (for which the salinities and calcium are known) we get the correct value with the Salifert.
Should not a standard, regardless of 'systemic' issues give the expected results?
We don't know in what matrix the Seachem "standard" is and also not how the documented value was obtained and what the error in that value is.