Another note on using the Reference solution:
Give yourself a range before you declare the probe is bad.
The reference solution is for verification that the probe is accurate.
1) Clean (soak) probe with special probe cleaning solution or vinegar (10-15 mins).
2) Rinse probe with clean water
3) remove excess water from probe
4) Insert probe into reference solution (see above post) and let it sit there for 5-10 mins
5) an exceptabe RANGE for 470 mv reference solution would be something like 450-490 mv (see bottle for exact range)
6)If probe is OUT of range..clean again and go back to step 1
7) If still out of range.....i dont know...get a new probe or look into calibrating the probe...you need other solutions to calibrate. I usually replace the probe after about 1-1.5 years anyhow.
The above is my opinion and is what has worked for me. If a scientist chimes in and says to do it another way...do it, quess.
Brad
Give yourself a range before you declare the probe is bad.
The reference solution is for verification that the probe is accurate.
1) Clean (soak) probe with special probe cleaning solution or vinegar (10-15 mins).
2) Rinse probe with clean water
3) remove excess water from probe
4) Insert probe into reference solution (see above post) and let it sit there for 5-10 mins
5) an exceptabe RANGE for 470 mv reference solution would be something like 450-490 mv (see bottle for exact range)
6)If probe is OUT of range..clean again and go back to step 1
7) If still out of range.....i dont know...get a new probe or look into calibrating the probe...you need other solutions to calibrate. I usually replace the probe after about 1-1.5 years anyhow.
The above is my opinion and is what has worked for me. If a scientist chimes in and says to do it another way...do it, quess.
Brad