Now it's my turn. My system:
two 12 cubic foot water softeners in series.
Three 9" KDF filters in parallell to strip chlorine.
Two-pass RO, 66% recovery, 20 BW30-4040 membranes, 19,000 GPD
3000 gallon storage tank, PVDF lined, Nitrogen blanket, .33 ppm ozone. All plumbing past this point is PVDF or stainless steel
20 gpm pump
Dual wavelength high-energy UV to kill ozone
3 pass Liqui-Cel degas membranes, combo mode (Vacuum + high purity N2 sweep)
2 E-Cell electrodeionization units in parallel
two 15 cubic foot mixed beds filled with Resintec LTOC resin
.45 micron filters
water goes back to the 3000 gallon tank, unless it needs to top of the pair of 2200 gallon tanks
two 2200 gallon tanks, PVDF lined, high purity nitrogen blanket, .033 ppm ozone
100 gpm pump
dual wavelength high energy UV to kill the ozone
three pass Liqui-cell degas membranes in sweep mode
two 30 cubic foot mixed beds, more Resintec LTOC resin
.45 micron filters
another 100 gpm pump
pair of ultrafilters, 6000 Dalton cutoff
Heat exchanger
system then feeds 22,000 square feet of clean room, and I forget how many square feet of lab space.
Back to the 2200 gallon tanks.
Our TOC level is typically below 200 PPT, Boron is typically below the detection limit of 15 ppt. Most semiconductor fabs are happy to get below 1 ppb of TOC. We only use virgin resin, what comes out of our system gets sold to people with less stringent standards.